I recently purchased an ADS1256 Module for $17.
The TI ADS1256 is a 24 Bit 8 Channel ADC with SPI communication. It offers a PGA with up to 64x gain and acquisition rates up to 30kHz.
It does not support negative voltages relative to ground for analog input as I learned (disappointingly).
Here is an image of the module:
I started working with the code developed by dariosalvi on the arduino forums here
I modified that code to get it working properly and added lots of comments and info from the datasheet.
I'm using it with a Teensy 3.2 running at 120 MHz and 2.5 MHz SPI
Currently compiling under Arduino 1.8.5 with Teensyduino 1.40
Here is the working code for reading all 8 single ended channels (AINx-AINCOM)
/* ADS1256, datasheet: http://www.ti.com/lit/ds/sbas288j/sbas288j.pdf
compare: https://github.com/Flydroid/ADS12xx-Library/blob/master/ads12xx.cpp
Reads all 8 Single Ended Channels AINx-AINCOM
Arduino 1.8.5, Teensyduino 1.40
Connections to Teensy 3.2
ADS1256 Teensy 3.2
DVDD (5V) - 5V
DGND (GND} - GND
SCLK - pin 13 (SCK)
DIN - pin 11 (MOSI)
DOUT - pin 12 (MISO)
DRDY - pin 9
CS - pin 10 (CS)
RESET - pin 8 (or tie HIGH?)
#define cs 10 // chip select
#define rdy 9 // data ready, input
#define rst 8 // may omit
#define SPISPEED 2500000 // Teensy 3.2 @120 mhz
pinMode(cs, OUTPUT);
digitalWrite(cs, LOW); // tied low is also OK.
pinMode(rdy, INPUT);
pinMode(rst, OUTPUT);
digitalWrite(rst, LOW);
delay(1); // LOW at least 4 clock cycles of onboard clock. 100 microseconds is enough
digitalWrite(rst, HIGH); // now reset to default values
SPI.begin(); //start the spi-bus
//init
while (digitalRead(rdy)) {} // wait for ready_line to go low
SPI.beginTransaction(SPISettings(SPISPEED, MSBFIRST, SPI_MODE1)); // start SPI
delayMicroseconds(100);
//Reset to Power-Up Values (FEh)
SPI.transfer(0xFE);
delay(5);
STATUS : STATUS REGISTER (ADDRESS 00h)
Reset Value = x1h
BIT 7 BIT 6 BIT 5 BIT 4 BIT 3 BIT 2 BIT 1 BIT 0
ID ID ID ID ORDER ACAL BUFEN DRDY
Bits 7-4 ID3, ID2, ID1, ID0 Factory Programmed Identification Bits (Read Only)
Bit 3 ORDER: Data Output Bit Order
0 = Most Significant Bit First (default)
1 = Least Significant Bit First
Input data is always shifted in most significant byte and bit first. Output data is always shifted out most significant
byte first. The ORDER bit only controls the bit order of the output data within the byte.
Bit 2 ACAL: Auto-Calibration
0 = Auto-Calibration Disabled (default)
1 = Auto-Calibration Enabled
When Auto-Calibration is enabled, self-calibration begins at the completion of the WREG command that changes
the PGA (bits 0-2 of ADCON register), DR (bits 7-0 in the DRATE register) or BUFEN (bit 1 in the STATUS register)
values.
Bit 1 BUFEN: Analog Input Buffer Enable
0 = Buffer Disabled (default)
1 = Buffer Enabled
Bit 0 DRDY: Data Ready (Read Only)
This bit duplicates the state of the DRDY pin.
byte status_reg = 0x00 ; // address (datasheet p. 30)
byte status_data = 0x01; // 01h = 0000 0 0 0 1 => status: Most Significant Bit First, Auto-Calibration Disabled, Analog Input Buffer Disabled
//byte status_data = 0x07; // 01h = 0000 0 1 1 1 => status: Most Significant Bit First, Auto-Calibration Enabled, Analog Input Buffer Enabled
SPI.transfer(0x50 | status_reg);
SPI.transfer(0x00); // 2nd command byte, write one register only
SPI.transfer(status_data); // write the databyte to the register
ADCON: A/D Control Register (Address 02h)
Reset Value = 20h
0 CLK1 CLK0 SDCS1 SDCS0 PGA2 PGA1 PGA0
Bit 7 Reserved, always 0 (Read Only)
Bits 6-5 CLK1, CLK0: D0/CLKOUT Clock Out Rate Setting
00 = Clock Out OFF
01 = Clock Out Frequency = fCLKIN (default)
10 = Clock Out Frequency = fCLKIN/2
11 = Clock Out Frequency = fCLKIN/4
When not using CLKOUT, it is recommended that it be turned off. These bits can only be reset using the RESET pin.
Bits 4-2 SDCS1, SCDS0: Sensor Detect Current Sources
00 = Sensor Detect OFF (default)
01 = Sensor Detect Current = 0.5μA
10 = Sensor Detect Current = 2μA
11 = Sensor Detect Current = 10μA
The Sensor Detect Current Sources can be activated to verify the integrity of an external sensor supplying a signal to the
ADS1255/6. A shorted sensor produces a very small signal while an open-circuit sensor produces a very large signal.
Bits 2-0 PGA2, PGA1, PGA0: Programmable Gain Amplifier Setting
PGA SETTING
00h = 000 = 1 ±5V(default)
01h = 001 = 2 ±2.5V
02h = 010 = 4 ±1.25V
03h = 011 = 8 ±0.625V
04h = 100 = 16 ±312.5mV
05h = 101 = 32 ±156.25mV
06h = 110 = 64 ±78.125mV
07h = 111 = 64 ±78.125mV
byte adcon_reg = 0x02; //A/D Control Register (Address 02h)
//byte adcon_data = 0x20; // 0 01 00 000 => Clock Out Frequency = fCLKIN, Sensor Detect OFF, gain 1
byte adcon_data = 0x00; // 0 00 00 000 => Clock Out = Off, Sensor Detect OFF, gain 1
//byte adcon_data = 0x01; // 0 00 00 001 => Clock Out = Off, Sensor Detect OFF, gain 2
SPI.transfer(0x50 | adcon_reg); // 52h = 0101 0010
SPI.transfer(adcon_data); // write the databyte to the register
DRATE: A/D Data Rate (Address 03h)
Reset Value = F0h
DR7 DR6 DR5 DR4 DR3 DR2 DR1 DR0
The 16 valid Data Rate settings are shown below. Make sure to select a valid setting as the invalid settings may produce
unpredictable results.
Bits 7-0 DR[7: 0]: Data Rate Setting(1)
F0h = 11110000 = 30,000SPS (default)
E0h = 11100000 = 15,000SPS
D0h = 11010000 = 7,500SPS
C0h = 11000000 = 3,750SPS
B0h = 10110000 = 2,000SPS
A1h = 10100001 = 1,000SPS
92h = 10010010 = 500SPS
82h = 10000010 = 100SPS
72h = 01110010 = 60SPS
63h = 01100011 = 50SPS
53h = 01010011 = 30SPS
43h = 01000011 = 25SPS
33h = 00110011 = 15SPS
23h = 00100011 = 10SPS
13h = 00010011 = 5SPS
03h = 00000011 = 2.5SPS
(1) for fCLKIN = 7.68MHz. Data rates scale linearly with fCLKIN.
byte drate_reg = 0x03; //DRATE: A/D Data Rate (Address 03h)
byte drate_data = 0xF0; // F0h = 11110000 = 30,000SPS
SPI.transfer(0x50 | drate_reg);
SPI.transfer(drate_data); // write the databyte to the register
// Perform Offset and Gain Self-Calibration (F0h)
SPI.transfer(0xF0);
delay(400);
digitalWrite(cs, HIGH);
SPI.endTransaction();
while (!Serial && (millis () <= 5000)); // WAIT UP TO 5000 MILLISECONDS FOR SERIAL OUTPUT CONSOLE
Serial.println("configured, starting");
Serial.println("");
Serial.println("AIN0-AINCOM AIN1-AINCOM AIN2-AINCOM AIN3-AINCOM AIN4-AINCOM AIN5-AINCOM AIN6-AINCOM AIN7-AINCOM");
//Single ended Measurements
unsigned long adc_val[8] = {0,0,0,0,0,0,0,0}; // store readings in array
byte mux[8] = {0x08,0x18,0x28,0x38,0x48,0x58,0x68,0x78};
int i = 0;
delayMicroseconds(2);
Settling Time Using the Input Multiplexer
The most efficient way to cycle through the inputs is to
change the multiplexer setting (using a WREG command
to the multiplexer register MUX) immediately after DRDY
goes low. Then, after changing the multiplexer, restart the
conversion process by issuing the SYNC and WAKEUP
commands, and retrieve the data with the RDATA
command. Changing the multiplexer before reading the
data allows the ADS1256 to start measuring the new input
channel sooner. Figure 19 demonstrates efficient input
cycling. There is no need to ignore or discard data while
cycling through the channels of the input multiplexer
because the ADS1256 fully settles before DRDY goes low,
indicating data is ready.
Step 1: When DRDY goes low, indicating that data is ready for retrieval,
update the multiplexer register MUX using the WREG command. For example,
setting MUX to 23h gives AINP = AIN2, AINN = AIN3.
Step 2: Restart the conversion process by issuing a SYNC command
immediately followed by a WAKEUP command.
Make sure to follow timing specification t11 between commands.
Step 3: Read the data from the previous conversion using the RDATA command.
Step 4: When DRDY goes low again, repeat the cycle by first
updating the multiplexer register, then reading the previous data.
for (i=0; i <= 7; i++){ // read all 8 Single Ended Channels AINx-AINCOM
byte channel = mux[i]; // analog in channels #
while (digitalRead(rdy)) {} ;
WREG: Write to Register
Description: Write to the registers starting with the register specified as part of the command. The number of registers that
will be written is one plus the value of the second byte in the command.
1st Command Byte: 0101 rrrr where rrrr is the address to the first register to be written.
2nd Command Byte: 0000 nnnn where nnnn is the number of bytes to be written – 1.
Data Byte(s): data to be written to the registers.
//byte data = (channel << 4) | (1 << 3); //AIN-channel and AINCOM // ********** Step 1 **********
//byte data = (channel << 4) | (1 << 1)| (1); //AIN-channel and AINCOM // ********** Step 1 **********
//Serial.println(channel,HEX);
SPI.transfer(0x50 | 0x01); // 1st Command Byte: 0101 0001 0001 = MUX register address 01h
SPI.transfer(0x00); // 2nd Command Byte: 0000 0000 1-1=0 write one byte only
SPI.transfer(channel); // Data Byte(s): xxxx 1000 write the databyte to the register(s)
//SYNC command 1111 1100 // ********** Step 2 **********
SPI.transfer(0xFC);
//while (!digitalRead(rdy)) {} ;
//WAKEUP 0000 0000
SPI.transfer(0x00);
delayMicroseconds(250); // Allow settling time
MUX : Input Multiplexer Control Register (Address 01h)
Reset Value = 01h
PSEL3 PSEL2 PSEL1 PSEL0 NSEL3 NSEL2 NSEL1 NSEL0
Bits 7-4 PSEL3, PSEL2, PSEL1, PSEL0: Positive Input Channel (AINP) Select
0000 = AIN0 (default)
0001 = AIN1
0010 = AIN2 (ADS1256 only)
0011 = AIN3 (ADS1256 only)
0100 = AIN4 (ADS1256 only)
0101 = AIN5 (ADS1256 only)
0110 = AIN6 (ADS1256 only)
0111 = AIN7 (ADS1256 only)
1xxx = AINCOM (when PSEL3 = 1, PSEL2, PSEL1, PSEL0 are “don’t care”)
NOTE: When using an ADS1255 make sure to only select the available inputs.
Bits 3-0 NSEL3, NSEL2, NSEL1, NSEL0: Negative Input Channel (AINN)Select
0000 = AIN0
0001 = AIN1 (default)
1xxx = AINCOM (when NSEL3 = 1, NSEL2, NSEL1, NSEL0 are “don’t care”)
SPI.transfer(0x01); // Read Data 0000 0001 (01h) // ********** Step 3 **********
delayMicroseconds(5);
adc_val[i] = SPI.transfer(0);
adc_val[i] <<= 8; //shift to left
adc_val[i] |= SPI.transfer(0);
adc_val[i] <<= 8;
} // Repeat for each channel ********** Step 4 **********
//The ADS1255/6 output 24 bits of data in Binary Two's
//Complement format. The LSB has a weight of
//2VREF/(PGA(223 − 1)). A positive full-scale input produces
//an output code of 7FFFFFh and the negative full-scale
//input produces an output code of 800000h.
for (i=0; i <= 7; i++){ // Single ended Measurements
if(adc_val[i] > 0x7fffff){ //if MSB == 1
adc_val[i] = adc_val[i]-16777216; //do 2's complement
Serial.print(adc_val[i]); // Raw ADC integer value +/- 23 bits
Serial.print(" ");
delay(250);
And here is the working code for reading all 4 differential channel pairs (AIN0-AIN1, AIN2-AIN3, AIN4-AIN5, AIN6-AIN7)
Reads all 4 differential channel pairs AIN0-AIN1, AIN2-AIN3, AIN4-AIN5, AIN6-AIN7
ADS1256 Teensy 3.2
DVDD (5V) - 5V
DGND (GND} - GND
SCLK - pin 13 (SCK)
DIN - pin 11 (MOSI)
DRDY - pin 9
CS - pin 10 (CS)
//init
SPI.transfer(0xFE);
delay(5);
Reset Value = x1h
values.
1 = Buffer Enabled
Reset Value = 20h
00 = Clock Out OFF
PGA SETTING
01h = 001 = 2 ±2.5V
Reset Value = F0h
SPI.transfer(0xF0);
delay(400);
Serial.println("");
Serial.println("AIN0-AIN1 AIN2-AIN3 AIN4-AIN5 AIN6-AIN7");
void loop(){
// Differential Measurements
long adc_val[4] = {0,0,0,0}; // store readings in array
byte mux[4] = {0x01,0x23,0x45,0x67};
int i = 0;
for (i=0; i <= 3; i++){ // read all 4 differential channel pairs AIN0-AIN1, AIN2-AIN3, AIN4-AIN5, AIN6-AIN7
SPI.transfer(0xFC);
//WAKEUP 0000 0000
SPI.transfer(0x00);
Reset Value = 01h
0001 = AIN1
0000 = AIN0
adc_val[i] <<= 8;
Serial.print(" ");
delay(250);
Next I want to try to get the Data Ready signal working correctly using interrupts on the Teensy.
All in all it's not a bad ADC module. The AINCOM on the module is tied to ground so it can't be used as a floating differential channel even though it's independently muxed to the ADC.
Also you can't use all 24 bits if you have a +/-5V signal without shifting and attenuating the signals. otherwise this module works very well and is well designed with seemingly high quality components.
There's also a library by Flydroid with many features for the ADS1256